Hello everyone,

I have looked at several datasets from Pier 14.

I have the same problem with all datasets. The flats have overcorrected.

So far I have always been able to solve this problem by doing a DBE.

Does anyone else have this problem, or am I doing something wrong?

 

CS

Darius

8 months ago
·
#6598
0
Votes
Undo

Hello everyone,

I have looked at several datasets from Pier 14.

I have the same problem with all datasets. The flats have overcorrected.

So far I have always been able to solve this problem by doing a DBE.

Does anyone else have this problem, or am I doing something wrong?

 

CS

Darius

 Hi Darius,

Yes we have been working on this issue and have managed to figure out the cause and a solution.

The cause is down to the gain settings we were trialling on Pier 14. We set them to the values suggested for Deepsky  by ZWO. It does work but then creates a much more noticeable variation in the bias pixel values,  the variation is inherent in CMOS cameras.....

The short answer there is pretty easy fix, this can be done in two ways by adding a pixel value to your Master Flats in Pixel Maths or when calibrating your flats before stacking to create your master,  simply  add a Pedestal value of 200 or above. Both methods will reduce the over correction of your light frames.

In truth the values are quite low and can be easily resolved with DBE as you suggested.

The long term fix it to set the Gain value back to our standard dynamic gain setting and will be implementing this in the coming weeks. We have paused the new jobs in the job queue  to allow the ones that have already started to finish, there are only a few jobs to complete now and once done we will change to the new gain setting and do some fresh calibration frames.

Hope this answers this for you.

 

regards

Peter

 

 


Peter Shah - Collimation & Telescope servicing.
Visit my personal imaging website at astropix.co.uk
For Image Processing Tutorials
Contact: pete@ccdimaging.co.uk


We can supply your new high quality Newtonian or Dall Kirkham Astrograph

Peter Shah
Roboscopes Observatory Controller


8 months ago
·
#6601
0
Votes
Undo

Thanks for clarifying Pete, I had noticed this on the narrowband frames from P2 & P9 also. 

I must confess I am quite partial to the idea of using slightly higher gain settings, particularly for the narrowband filters and particularly on the f/6.8 systems...  The more efficient unguided 180s frames are what I think we all want to be using, and perhaps these might be best served by pushing the gain a bit for certain systems/applications?  

Of course such ideas stem from my upbringing as a backyard imager, so feel free to dismiss as ridiculous :)

8 months ago
·
#6602
0
Votes
Undo

Thanks for clarifying Pete, I had noticed this on the narrowband frames from P2 & P9 also. 

I must confess I am quite partial to the idea of using slightly higher gain settings, particularly for the narrowband filters and particularly on the f/6.8 systems...  The more efficient unguided 180s frames are what I think we all want to be using, and perhaps these might be best served by pushing the gain a bit for certain systems/applications?  

Of course such ideas stem from my upbringing as a backyard imager, so feel free to dismiss as ridiculous :)

you will see it on every pier if you do not bias remove your flats before creating your master...here is an example from pier 1.

 

Higher gain setting is much better as the S/N is much more favourable. Manuel is doing a lot of work on this subject and should have some very interesting result coming soon. Higher gain and shorter sub lengths seem to yield high signal in the data sets...from what I am seeing 300s is actually too long and you will see more depth after a stack with the same amount of 120s  ......I cant wait to see Manuel's full conclusions on this...

 

cheers Peter


Peter Shah - Collimation & Telescope servicing.
Visit my personal imaging website at astropix.co.uk
For Image Processing Tutorials
Contact: pete@ccdimaging.co.uk


We can supply your new high quality Newtonian or Dall Kirkham Astrograph

Peter Shah
Roboscopes Observatory Controller


0
Votes
Undo

Hello everyone,

I have looked at several datasets from Pier 14.

I have the same problem with all datasets. The flats have overcorrected.

So far I have always been able to solve this problem by doing a DBE.

Does anyone else have this problem, or am I doing something wrong?

 

CS

Darius

 Hi Darius,

Yes we have been working on this issue and have managed to figure out the cause and a solution.

The cause is down to the gain settings we were trialling on Pier 14. We set them to the values suggested for Deepsky  by ZWO. It does work but then creates a much more noticeable variation in the bias pixel values,  the variation is inherent in CMOS cameras.....

The short answer there is pretty easy fix, this can be done in two ways by adding a pixel value to your Master Flats in Pixel Maths or when calibrating your flats before stacking to create your master,  simply  add a Pedestal value of 200 or above. Both methods will reduce the over correction of your light frames.

In truth the values are quite low and can be easily resolved with DBE as you suggested.

The long term fix it to set the Gain value back to our standard dynamic gain setting and will be implementing this in the coming weeks. We have paused the new jobs in the job queue  to allow the ones that have already started to finish, there are only a few jobs to complete now and once done we will change to the new gain setting and do some fresh calibration frames.

Hope this answers this for you.

 

regards

Peter

 

 

Thanks for clarifying Peter. The method with Pixelmath also worked well.

8 months ago
·
#6604
0
Votes
Undo

Hello everyone,

I have looked at several datasets from Pier 14.

I have the same problem with all datasets. The flats have overcorrected.

So far I have always been able to solve this problem by doing a DBE.

Does anyone else have this problem, or am I doing something wrong?

 

CS

Darius

 Hi Darius,

Yes we have been working on this issue and have managed to figure out the cause and a solution.

The cause is down to the gain settings we were trialling on Pier 14. We set them to the values suggested for Deepsky  by ZWO. It does work but then creates a much more noticeable variation in the bias pixel values,  the variation is inherent in CMOS cameras.....

The short answer there is pretty easy fix, this can be done in two ways by adding a pixel value to your Master Flats in Pixel Maths or when calibrating your flats before stacking to create your master,  simply  add a Pedestal value of 200 or above. Both methods will reduce the over correction of your light frames.

In truth the values are quite low and can be easily resolved with DBE as you suggested.

The long term fix it to set the Gain value back to our standard dynamic gain setting and will be implementing this in the coming weeks. We have paused the new jobs in the job queue  to allow the ones that have already started to finish, there are only a few jobs to complete now and once done we will change to the new gain setting and do some fresh calibration frames.

Hope this answers this for you.

 

regards

Peter

 

 

Thanks for clarifying Peter. The method with Pixelmath also worked well.

No problem Darius...


Peter Shah - Collimation & Telescope servicing.
Visit my personal imaging website at astropix.co.uk
For Image Processing Tutorials
Contact: pete@ccdimaging.co.uk


We can supply your new high quality Newtonian or Dall Kirkham Astrograph

Peter Shah
Roboscopes Observatory Controller


8 months ago
·
#6605
0
Votes
Undo

Hi all, this is the reason we have not inputted new jobs into P14 yet, we want to alter the gain first. There are only 2 jobs left to finish now.

Once done we will let you know and start getting new darks and flats

 

Steve


Please ignore my dylexia wherever possible, just be thankful I can control my Tourettes ;)

Things to do, so little time!

Steve
Roboscopes Tea Boy


8 months ago
·
#6606
0
Votes
Undo

Hey

 

Thanks for this I started looking at p14  stuff a couple opf days ago and i thought that i was going mad!  was planning a 

set of integration methodsd for the flats to see if that had any effect  - this ,may be a lovely quick way aof sorthing this out........

8 months ago
·
#6611
0
Votes
Undo

Hello everyone,

I have looked at several datasets from Pier 14.

I have the same problem with all datasets. The flats have overcorrected.

So far I have always been able to solve this problem by doing a DBE.

Does anyone else have this problem, or am I doing something wrong?

 

CS

Darius

 Hi Darius,

Yes we have been working on this issue and have managed to figure out the cause and a solution.

The cause is down to the gain settings we were trialling on Pier 14. We set them to the values suggested for Deepsky  by ZWO. It does work but then creates a much more noticeable variation in the bias pixel values,  the variation is inherent in CMOS cameras.....

The short answer there is pretty easy fix, this can be done in two ways by adding a pixel value to your Master Flats in Pixel Maths or when calibrating your flats before stacking to create your master,  simply  add a Pedestal value of 200 or above. Both methods will reduce the over correction of your light frames.

In truth the values are quite low and can be easily resolved with DBE as you suggested.

The long term fix it to set the Gain value back to our standard dynamic gain setting and will be implementing this in the coming weeks. We have paused the new jobs in the job queue  to allow the ones that have already started to finish, there are only a few jobs to complete now and once done we will change to the new gain setting and do some fresh calibration frames.

Hope this answers this for you.

 

regards

Peter

 

 

Thanks for clarifying Peter. The method with Pixelmath also worked well.

 

 

Would you be able to share what you did in pixel math?


“There are no bad pictures; that's just how your face looks sometimes.”

― Abraham Lincoln


0
Votes
Undo

Hello everyone,

I have looked at several datasets from Pier 14.

I have the same problem with all datasets. The flats have overcorrected.

So far I have always been able to solve this problem by doing a DBE.

Does anyone else have this problem, or am I doing something wrong?

 

CS

Darius

 Hi Darius,

Yes we have been working on this issue and have managed to figure out the cause and a solution.

The cause is down to the gain settings we were trialling on Pier 14. We set them to the values suggested for Deepsky  by ZWO. It does work but then creates a much more noticeable variation in the bias pixel values,  the variation is inherent in CMOS cameras.....

The short answer there is pretty easy fix, this can be done in two ways by adding a pixel value to your Master Flats in Pixel Maths or when calibrating your flats before stacking to create your master,  simply  add a Pedestal value of 200 or above. Both methods will reduce the over correction of your light frames.

In truth the values are quite low and can be easily resolved with DBE as you suggested.

The long term fix it to set the Gain value back to our standard dynamic gain setting and will be implementing this in the coming weeks. We have paused the new jobs in the job queue  to allow the ones that have already started to finish, there are only a few jobs to complete now and once done we will change to the new gain setting and do some fresh calibration frames.

Hope this answers this for you.

 

regards

Peter

 

 

Thanks for clarifying Peter. The method with Pixelmath also worked well.

 

 

Would you be able to share what you did in pixel math?

I loaded my MasterFlat into Pixinsight and added pixel values to it with Pixelmath.

 

Attachments (1)
8 months ago
·
#6613
0
Votes
Undo

Hello everyone,

I have looked at several datasets from Pier 14.

I have the same problem with all datasets. The flats have overcorrected.

So far I have always been able to solve this problem by doing a DBE.

Does anyone else have this problem, or am I doing something wrong?

 

CS

Darius

 Hi Darius,

Yes we have been working on this issue and have managed to figure out the cause and a solution.

The cause is down to the gain settings we were trialling on Pier 14. We set them to the values suggested for Deepsky  by ZWO. It does work but then creates a much more noticeable variation in the bias pixel values,  the variation is inherent in CMOS cameras.....

The short answer there is pretty easy fix, this can be done in two ways by adding a pixel value to your Master Flats in Pixel Maths or when calibrating your flats before stacking to create your master,  simply  add a Pedestal value of 200 or above. Both methods will reduce the over correction of your light frames.

In truth the values are quite low and can be easily resolved with DBE as you suggested.

The long term fix it to set the Gain value back to our standard dynamic gain setting and will be implementing this in the coming weeks. We have paused the new jobs in the job queue  to allow the ones that have already started to finish, there are only a few jobs to complete now and once done we will change to the new gain setting and do some fresh calibration frames.

Hope this answers this for you.

 

regards

Peter

 

 

Thanks for clarifying Peter. The method with Pixelmath also worked well.

 

 

Would you be able to share what you did in pixel math?

I loaded my MasterFlat into Pixinsight and added pixel values to it with Pixelmath.

 

So adding 0.03 was enough, without a pedestal for example? I'll give it a go, thanks.


“There are no bad pictures; that's just how your face looks sometimes.”

― Abraham Lincoln


8 months ago
·
#6614
0
Votes
Undo

I'm not having much success with this! I tried various values and used a single blue frame and tried to correct it with bias/dark/flat - these are the results.

 

https://imgur.com/a/dNJH7e7


“There are no bad pictures; that's just how your face looks sometimes.”

― Abraham Lincoln


0
Votes
Undo

I'm not having much success with this! I tried various values and used a single blue frame and tried to correct it with bias/dark/flat - these are the results.

 

https://imgur.com/a/dNJH7e7

That is strange. In pixinsight I also have problems to calibrate the lights with the masterflat. So far I have only used Pixinsight to process the masterflat with PixelMath. I calibrated and stacked the lights in AstroPixelProcessor. There the results look ok. This is a single blue frame after calibration. Here I had added a value of 0.10 to the masterflat. I am still experimenting with the exact values. The 0.03 I had used for a narrowband channel. These seem to be less affected than the broadband channels. I hope this helps.

CS
Darius

8 months ago
·
#6617
0
Votes
Undo

I am going to try making a truly synthetic flat in photoshop, based up on an actual light frame... lets see how that goes. Will post results!


“There are no bad pictures; that's just how your face looks sometimes.”

― Abraham Lincoln


8 months ago
·
#6619
0
Votes
Undo

To continue my monologue... no point posting results LOL!


“There are no bad pictures; that's just how your face looks sometimes.”

― Abraham Lincoln


8 months ago
·
#6651
0
Votes
Undo

I find this method of making synthetic flats is much easier then trying to make a synthetic flat in Photo Shop or Affinity Photo.

Synthetic Flat

With my QHY-268c and QHY-183 I find dark flats work much better then bias frames. This way dark flat gain and exposure time always match the flats. With this I never really need to use flat or dark optimization. I had to get creative dealing with the QHY 183's insane amp glow and dark flats and no bias was the winner.

Try some dark flats for pier 14 :-).

Daniel

8 months ago
·
#6660
0
Votes
Undo
Thanks Daniel, that is a useful link! I have also switched to using dark flats for my backyard setup - I highly recommend we investigate using these for our CMOS cams...
  • Page :
  • 1
There are no replies made for this post yet.
Be one of the first to reply to this post!

Follow Us

Newsletter

Proud to use

  • FLI

  • 656 Imaging

  • 10 Micron

  • Planewave

  • ZWO

Company Details:

Roboscopes

802 Kingsbury Road
Birmingham
B24 9PS
United Kingdom


Roboscopes is a trading name of ENS Optical LTD ¦ Copyright© 2020 Roboscopes
Cron Job Starts